I have the rev ed retailed Honeywell sensor and it is giving me up to 122% Humidity readings. Power is 3 x AA cells, actually giving 4.92V.
Here are my calcs, I'm sure they are OK but at full 10 bit ADC (1023) it will allow the sensor to indicate 135% RH. Any one else have this problem?
VOut = Vsupply ((0.0062 x SensorRH) + 0.16 )
Therefore VOut/Vsupply = (0.0062 x SensorRH) + 0.16 = ADC/1023
So, ADC/1023 = (0.0062 x SensorRH) + 0.16, dividing thru by 0.0062
SensorRH = (ADC/1023) x (1/0.0062) - (0.16/0.0062)
Therefore SensorRH = (ADC x 0.1576) - 25.80646
Or for code for the PicAxe :
63xSensorRH = 10 x ADC - 1665
The sensor seems right, at midday it was showing 66% humidity, (local airport 30 Kms away near the sea was showing 73% on the net).
This increased to 98% at dusk, then this morning it was 122%. Weather was foggy, so was definitely 100% RH.
I'm sure the sensor is Ok so I might need to limit the 100% reading in software. Just wondered if there are any similar experiences.
Dowmload cable is not connected, so it's not that skewing the number.
Here are my calcs, I'm sure they are OK but at full 10 bit ADC (1023) it will allow the sensor to indicate 135% RH. Any one else have this problem?
VOut = Vsupply ((0.0062 x SensorRH) + 0.16 )
Therefore VOut/Vsupply = (0.0062 x SensorRH) + 0.16 = ADC/1023
So, ADC/1023 = (0.0062 x SensorRH) + 0.16, dividing thru by 0.0062
SensorRH = (ADC/1023) x (1/0.0062) - (0.16/0.0062)
Therefore SensorRH = (ADC x 0.1576) - 25.80646
Or for code for the PicAxe :
63xSensorRH = 10 x ADC - 1665
The sensor seems right, at midday it was showing 66% humidity, (local airport 30 Kms away near the sea was showing 73% on the net).
This increased to 98% at dusk, then this morning it was 122%. Weather was foggy, so was definitely 100% RH.
I'm sure the sensor is Ok so I might need to limit the 100% reading in software. Just wondered if there are any similar experiences.
Dowmload cable is not connected, so it's not that skewing the number.